Westonci.ca is the Q&A platform that connects you with experts who provide accurate and detailed answers. Discover detailed solutions to your questions from a wide network of experts on our comprehensive Q&A platform. Join our platform to connect with experts ready to provide precise answers to your questions in different areas.

Low concentrations of EDTA near the detection limit gave the following dimensionless instrument readings: 175, 104, 164, 193, 131, 189, 155, 133, 151, and 176. Ten unknowns have a mean reading of 60.0. The slope of the calibration curve is 1.75×109M−1.
a) Estimate the signal detection limit for EDTA.
b) What is the concentration detection limit?
c) What is the lower limit of quantitation?


Sagot :

Answer:

Following are the solution to these question:

Explanation:

Calculating the mean:

[tex]\bar{x}=\frac{175+104+164+193+131+189+155+133+151+176}{10}\\\\[/tex]

  [tex]=\frac{1571}{10}\\\\=157.1[/tex]

Calculating the standardn:

[tex]\sigma=\sqrt{\frac{\Sigma(x_i-\bar{x})^2}{n-1}}\\\\[/tex]

Please find the correct equation in the attached file.

[tex]=28.195[/tex]

For point a:

[tex]=3s+yblank \\\\=3 \times 28.195+50\\\\=84.585+50\\\\=134.585\\[/tex]

For point b:

[tex]=3 \ \frac{s}{m}\\\\ = \frac{(3 \times 28.195)}{1.75 \times 10^9 \ M^{-1}}\\\\= 4.833 \times 10^{-8} \ M[/tex]

For point c:

[tex]= 10 \frac{s}{m} \\\\= \frac{(10 \times 28.195)}{1.75 x 10^9 \ M^{-1}}\\\\ = 1.611 \times 10^{-7}\ M[/tex]

It is calculated by using the slope value that is [tex]1.75 \times 10^9 M^{-1}[/tex]. The slope value [tex]1.75 \times 10^9 M^{-1}[/tex]is ambiguous.