Welcome to Westonci.ca, your ultimate destination for finding answers to a wide range of questions from experts. Discover in-depth answers to your questions from a wide network of experts on our user-friendly Q&A platform. Join our Q&A platform to connect with experts dedicated to providing accurate answers to your questions in various fields.

Consider a process that includes careful testing of each manufactured defibrillator. Listed ( below are the numbers of defective defibrillators in 20 successive batches of 10,000. Find the values you would use for the centerline, the upper control limit, and the lower control limit.
20 14 22 27 12 12 18 23 25 19 24 28 21 25 17 19 17 22 15 20
a. p = 0.040000
UCL 0.041 039, LCL = 0.03 8961
b. p 0.002000
UCL= 0.003340, LCL= 0.000660
c. p=0.0020000
UCL = 0, 021200, LCL = 0
d. p 0.040000
UCL = 0.041470, LCL = 0.038530


Sagot :

Your correct answer is C.
Thank you for your visit. We are dedicated to helping you find the information you need, whenever you need it. We hope you found what you were looking for. Feel free to revisit us for more answers and updated information. Stay curious and keep coming back to Westonci.ca for answers to all your burning questions.