Looking for answers? Westonci.ca is your go-to Q&A platform, offering quick, trustworthy responses from a community of experts. Connect with a community of experts ready to provide precise solutions to your questions on our user-friendly Q&A platform. Connect with a community of professionals ready to provide precise solutions to your questions quickly and accurately.
Sagot :
Answer:
b. 12.5 mAs, 70 kVp
Explanation:
The given parameter are;
The initial exposure factors := 10 mAs and 70 kVp
The initial Grid Ratio, G.R.₁ = 8:1
The Grid Ratio with which the radiographer desires to increase the scatter absorption, G.R.₂ = 12:1
Given that the lead content in the 12:1 grid, is higher than the lead content in 8:1 grid and that 12:1 grid needs more mAs to compensate, and provides a higher image contrast, the amount of extra mAs is given by the Grid Conversion Factors, GCF, as follows;
The GCF for G.R. 8:1 = 4
The GCF for G.R. 12:1 = 5
Therefore, given that the mAs used by the radiographer for 8:1 Grid Ratio is 10 mAs, the mAs required for a G.R. of 12:1 in order to maintain the same exposure is given as follows;
mAs for G.R. of 12:1 = 10 mAs × 5/4 = 12.5 mAs
Therefore the new exposure factors are;
12.5 mAs, 70 kVp
We hope our answers were helpful. Return anytime for more information and answers to any other questions you may have. Thanks for stopping by. We strive to provide the best answers for all your questions. See you again soon. Thank you for visiting Westonci.ca, your go-to source for reliable answers. Come back soon for more expert insights.