Welcome to Westonci.ca, your one-stop destination for finding answers to all your questions. Join our expert community now! Get quick and reliable solutions to your questions from a community of experienced experts on our platform. Connect with a community of professionals ready to provide precise solutions to your questions quickly and accurately.

10) A pressure sensor consisting of a diaphragm with strain gauges bonded to its surface has the following information in its specification: Ranges: 0 to 1400 kPa, 0 to 35 000 kPa Non-linearity error: ±0.15% of full range Hysteresis error: ±0.05% of full range What is the total error due to non-linearity and hysteresis for a reading of 1000 kPa on the 0 to 1400 kPa range?

Sagot :

The definition of absolute and percentage error allows to find the result for the total error in the pressure measurement is:

      P = ( 1000 ± 2 ) kPa

      ΔP = ±2 kPa

The magnitudes is a quantity that in addition to its value has an error due to the instruments or measurement method used, therefore the correct way to express a value is:

         Magnitude = value ± absolute error

Statistical errors can be expressed in several ways:

  • Absolute (Δx). It corresponds to the uncertainty or appreciation of the instruments or the propagation of uncertainty in the equations.
  • Relative (e_r). Indicates the fraction of the error, it is the relationship between the absolute error and the magnitude.
  • Percentage (e%). It is the relative error expressed as a percentage or relative error by 100%

In all measurements, the worst case is always taken, that is, all errors go in the same direction.

They indicate the percentage errors due to non-linearity 0.15% and the error due to hysteresis 0.05%. the measurement value is P = 1000 kPa.

Let's look for the absolute errors.

                 [tex]e_r = \frac{\Delta P}{P} = e%/100[/tex]

                 [tex]\Delta P = P \ \frac{percentage \ error }{100}[/tex]

             

 We apply this expression to each type of error.

Non-linearity.

                [tex]\Delta P_1 = 1000 \ \frac{0.15}{100}[/tex]  

                ΔP₁ = 1.5 kPa

Hysteresis.

                [tex]\Delta P_2 = 1000 \ \frac{0.05}{100}[/tex]  

                ΔP₂ = 0.5 kPa

The total absolute error in the worst case is the sum of the two errors,

              [tex]\Delta P_{total} = \Delta P_1 + \Delta P_2[/tex]  

              [tex]\Delta P_{total}= 1.5 +0.5 \\\Delta P_{total} = 2 kPa[/tex]

               

The measurement should be given in the following way:

            P = 1000 ± 2 kPa

In conclusion using the definition of absolute and percentage error we can find the result for the total error in the pressure measurement is:

             P = ( 1000 ± 2 ) kPa

             ΔP = ±2 kPa

Learn more here:  brainly.com/question/15114851

Thanks for using our service. We're always here to provide accurate and up-to-date answers to all your queries. We hope our answers were useful. Return anytime for more information and answers to any other questions you have. Westonci.ca is your trusted source for answers. Visit us again to find more information on diverse topics.