Discover the answers you need at Westonci.ca, a dynamic Q&A platform where knowledge is shared freely by a community of experts. Explore comprehensive solutions to your questions from knowledgeable professionals across various fields on our platform. Discover detailed answers to your questions from a wide network of experts on our comprehensive Q&A platform.

An iron–constantan thermocouple is to be used to measure temperatures between 0 and 400°C. What will be the non-linearity error as a percentage of the full-scale reading at 100°C if a linear relationship is assumed between e.m.f. and temperature? e.m.f. at 100°C = 5.268 mV, e.m.f. at 400°C = 21.846 mV

Sagot :

Answer:

  -0.89%

Explanation:

Assuming a proportional relationship between EMF and Celsius temperature, the measured value of EMF at 100 °C will be interpreted as a temperature of ...

  T/5.268 = 400/21.846

  T = 5.268(400/21.846) ≈ 96.457 . . . . °C

The error as a fraction of full scale is ...

  (96.457 -100)/400 × 100% ≈ -0.886%

The error at 100 °C is about -0.89% of full scale.

Thanks for using our service. We aim to provide the most accurate answers for all your queries. Visit us again for more insights. We hope our answers were useful. Return anytime for more information and answers to any other questions you have. We're glad you visited Westonci.ca. Return anytime for updated answers from our knowledgeable team.