Get the answers you need at Westonci.ca, where our expert community is dedicated to providing you with accurate information. Join our platform to connect with experts ready to provide precise answers to your questions in various areas. Get detailed and accurate answers to your questions from a dedicated community of experts on our Q&A platform.

An iron–constantan thermocouple is to be used to measure temperatures between 0 and 400°C. What will be the non-linearity error as a percentage of the full-scale reading at 100°C if a linear relationship is assumed between e.m.f. and temperature? e.m.f. at 100°C = 5.268 mV, e.m.f. at 400°C = 21.846 mV

Sagot :

Answer:

  -0.89%

Explanation:

Assuming a proportional relationship between EMF and Celsius temperature, the measured value of EMF at 100 °C will be interpreted as a temperature of ...

  T/5.268 = 400/21.846

  T = 5.268(400/21.846) ≈ 96.457 . . . . °C

The error as a fraction of full scale is ...

  (96.457 -100)/400 × 100% ≈ -0.886%

The error at 100 °C is about -0.89% of full scale.